Hm. PUT stands for "power up test", which is basic diagnostics stuff for the Sequoia. It looks like 25 and 28 are both Transmitter tests, in board slot 1. Since it is a C256, it only has one TX and one BF (Transmitter and Beamformer).
PUT 25 - Test TX VGain Slot 1 - Ensures TX VGAIN is functional for Slot 1, Fatal Error: system will try to boot -- possible causes: TX, CN (Controller)
PUT 28 - Test TX PWG registers Slot 1 - Safety, patient exposure to uncontrolled power levels. Fatal Error: system will try to boot -- possible causes: TX, CN
There's a few problems with this though. Unless you have "Third Party Service" turned on, you cannot pull the side cover off to do any sort of board maintenance/replacing (if you do pull the silver panel/cover off, the system software prevents the Sequoia from booting to imaging, and it puts the fans on really high/fast). I was going to recommend pulling out the TX and putting it back in to see if there's a loose/weak connection. However, I would need you to enter the "Feature Control" of the system, from the Service Tool, and look for the third party service option.
If you have it, great. I'd recommend pulling the side off, and reseating the TX and CN boards. Unless the local service rep has parts, that's the cheapest/easiest way to see if maybe if things were knocked loose in shipping.
If you don't have Third Party Service, you'll need to get a service password for the day. This will also allow you to run diagnostic tests to get a clearer picture of what could be wrong with the system.
Universal Diagnostic Solutions