Topic: Todd Help PLease

Hi
I don't know where I have to start
but this is the story

I have Acuson 128XP / 10C

I tried to fix this problem

*****************************************
Softwate sono-pac for right
transducer has not been licensed
Please remove right Transducer
*****************************************

1 - I had Changed batteries
2 - I made a test and I get this result
*** PM completed ,Failed 2 test
$
$Stat
F0 SCNR MXK TEST
226 OCR Interrupt Test
$

Please I need the encryption settings or ( strings ) for Acuson 128xp/10 NO:14525
can you please E-mailed encryption settings or ( strings ) to me
mirage4pc@yahoo.com

Re: Todd Help PLease

Youssef,

Please check your email when you have a moment.

As for the F0 SCNR MXK and OC3 RTC Interrupt tests, if you're still able to boot to imaging I wouldn't worry a whole lot about these until you're able to start scanning again.

The MXK test failure could be a false failure, or one of the scanner boards could be defective.

OC3 RTC (Real Time Clock) Interrupt test - there isn't a whole lot of information on this error code in the manual, but it does say that the two possible causes of this error are the OC3 and CVC. All I see in the manual for the CVC clock is "Board contains Time of Day Clock and BBM (battery backed memory)". My guess is the time-keeper component is faulty somehow?

At some point this OC3 error may lock the system up intermittently, and the MXK issue may cause some problems with the image, but if you're fine with that.. then just keep using it.

Universal Diagnostic Solutions