Topic: Aspen system error during power-on diagnostic

System model: Apsen advanced

When I first time turned on system power, the power-on diagnostic process was running step by step.
About 5 minutes later, the process stopped and an message appeared in display:
" PONLOG ERRORS:
SCP    0677
SVC    0626
RP      0434
RP      0433
RP      0427"

and then became:
"A SYSTEM ERROR HAS OCCURRED. PLEASE RESET SYSTEM."

There is also a smell.
No digit is shown on rear-end Power Supply Error LCD.

What happened?

Re: Aspen system error during power-on diagnostic

Well.. I can give you an idea as to what each of those error codes means, but as for what happened to cause the issue.. I can't be certain. The usual questions come to mind though: are you using a power conditioner, as well as a step-down transformer? If so, what wattage is the transformer rated for? Have you had any power surges lately, have the circuit breakers ever been tripped due?

In regards to the error codes, here's a brief breakdown:
(Scanner  Control Processor) SCP 0677 -- points to the PPS (Programmable Power Supply), it's a voltage accuracy test. If this is failing the PPS may not be outputting the correct voltages for boot-up.

(SYSCON & Video Controller) SVC 0626 -- points to the CDM (Cine Data Memory) and SVC. Probably an issue with communication, perhaps a power issue again, or a reference clock issue with the RP.

(Reconstruction Processor) RP 0434 --  points to the WFP (Waveform Processor) and RP boards.

RP 0433 -- points to the AQP (Acquisition Processor) and RP boards. Says it's an interface test, which I'm assuming means the AQP and RP aren't communicating properly. Could be a power issue to the boards, or a defective component on the RP.

RP 0427 -- points to the RP and SVC. This seems like a rather nasty failure. Again though... it could be a power problem with the main supply, since this board requires +5v.

I would recommend checking the voltages on the main power supply (right side of the system when facing the monitor, behind the side panel).

What you want to look for is: 5v, 7.5v, 12v and 15v, there is also a 300vdc output voltage.

ALL scan converter boards operate on the +5v digital rail. Some of them use +12v, including the AQP, CDM and SVC. Check both of those and let me know.

Universal Diagnostic Solutions

3

Re: Aspen system error during power-on diagnostic

Todd, thanks for your reply.

There are three parts in Aspen power module:
an MDI 2164, a PPS3 box, and an OVP box.
I just find the +5V & RTN labels on MDI 2164, but I cannot find any other labels for power supply connectors.
I have no service manual of this system.
Could you tell me where are the 7.5V, 12V, 15V and 300V?
Thanks again.

4

Re: Aspen system error during power-on diagnostic

I have measured MDI PS 2164.
+12V is measured at J8;
+15V, +7.5V, and +5V are measured at J7B;
-15V and -7.5V are measured at J7A.

I've also measured J3 of PPS box,
+301V, +10.43V, 54.3V(ac), and 8.0V(ac) are measured.

At OVP box, +10.39V is measured.

Re: Aspen system error during power-on diagnostic

Sorry for taking so long to reply back.

It looks like all of your main voltages for scanner/scan converter are good. This would lead me to believe that there is some sort of board fault.

My best guess would be the RP board. Since it controls such a large portion of bootup, and if it's failing, it could cause potential failures down the chain.

Do you have access to spare parts for the Aspen?

Universal Diagnostic Solutions

6

Re: Aspen system error during power-on diagnostic

I have a fundamental question: Does Aspen accept 110~220Vac line voltage?

The system is plugged into 110Vac line voltage for the 1st-time power on.
I have waited for about half minute until there was a smell.
And then I shutted it down and unplugged the power.

The system was pulgged into 220Vac for the 2nd-time power on,
and the result is described as my 2011-10-28 post.

Last week I pulgged it into 110Vac again and the final result is as follows,
Monitor: no power, no display;
Panel: it seems function well.
Printed out picture: Only text, symbol and dots. No ultrasound image.

Some measured result is listed for reference.
(1) Line voltage = 220Vac,
Voltage of J20 of MDI PS 2164 is 218Vac;
Voltages of J3 of PPS box,
+301V, +10.43V, 54.3V(ac), and 8.0V(ac).

(2) Line voltage = 110Vac,
Voltage of J20 of MDI PS 2164 is 23.7Vac;
Voltages of J3 of PPS box,
+301V, +8.20V, 28V(ac), and 27V(ac).

Any comment? Power supply damaged? PCB (RP or .... ) fault?

Re: Aspen system error during power-on diagnostic

The Aspen was designed to accept 110v only. However, field replaceable units (isolation transformer, AC input) were made to allow the system to run off of 220v, but not 110v.

Universal Diagnostic Solutions