Hello,

I have a problem with HDI 5000. It has an error code :

An Error has detected
Please record the following information and
call ATL Technical Support

Date: 04-28-2009 Time:17:30
Diagnostics Number: 0090

When this error appear, XRes key don't works

Thanks and best regards

Re: Error 90

I'll look into this error code today and see if I can nail it down for you.

If the XRes key isn't working though, I'd venture a guess that the APM device isn't connecting to the ultrasound. On the back of the system there should be a rectangular black box that connects to the system with a BNC on the other side that routes up into the APM. If that gets knocked loose or moved around you tend to lose connection to the device attached, as well as damage the CPU board potentially.

I'd look there first and then check to see if the APM is booting up when the machine is turned on.

Universal Diagnostic Solutions

Re: Error 90

Hello,

Thanks for your quick repply. I see APM module and this is booting up with the machine. When the transceiver lose connection, error code is 0089, not 0090.
Can you send me a list for error codes ? I have a copy of service manual, but dont have error codes... Thanks for your cooperation

Re: Error 90

I do not have access to that proprietary information from Philips, but I was able to reach a field technician. The error code 0089 is: No DVS bootup status. If other HDI functionality is working, either the HDI has a problem communicating with the DVS (failure in CPU, the ethernet cable, or the system software), or DVS did not bootup (failure in the DVS box).

0090 is basically the same thing.

Hope this helps.