Topic: Setting up Paddle Switches -- Depth, Log Comp, Trans Zone

Depending on the type of faceplate you have, the options for changing the depth, log compression, and transmit zone are located in the CSR menu.

As explained in previous posts, to reach this you need to press the Code+Shift+V+Test keys simultaneously. Once in the menu, scroll down to "Setup Paddle Switches" and Select it using the soft key on the monitor.

Xmit Zone Configuration

128/XP Radiology Configuration  ON/OFF
128 Radiology Configuration       ON/OFF

MultiHertz Control                PADDLE/KEY

The ON/OFF will have one or the other highlighted depending on what configuration you currently have set up. The paddle/key for the MultiHertz control is probably going to be dictated by your keyboard setup. If you have the MultiHtz Key on the right side, set it to KEY. If you have the MultiHtz paddle switch/toggle towards the middle of the user interface, change it to PADDLE.

You can set both the 128/XP Rad Config, and 128 Rad Config to OFF. This changes the configuration to the newest version of the faceplate that has "Log Comp/Depth/Trans Zone" labeling.

Universal Diagnostic Solutions

Re: Setting up Paddle Switches -- Depth, Log Comp, Trans Zone

Todd, Thanks for the explanation and instructions. I can handle it from here. I appreciate that you offer to help out on this forum.  Tim